2 current measurements with one 5740A meter

Post by don0101 »

I have 2 AC currents to be measured. I intend to use a 100A:5A AC current shunt on each wire. Can I use a Double Throw switch to switch one 5740A meter between the 2 shunts without causing damage to the shunt that is not connected to the meter? Thanks.
Tech_Marco
Posts: 4647
Joined: Mon Jun 15, 2009 3:00 pm
Contact:

Re: 2 current measurements with one 5740A meter

Post by Tech_Marco »

Yes you can. It won't damage the C.T. (current transformer) but both C.T. must be matched in size