No tickle pulse present - LO-X7

Post Reply
ian4goats
Posts: 17
Joined: Wed Apr 09, 2014 10:26 am
Location: Danvers, MA
Contact:

No tickle pulse present - LO-X7

Post by ian4goats »

Hello,

We have recently discontinued using Access laser tubes and have moved to Synrad due to mode hop issues.
The Access tube did not need pre-ignition or tickle pulse.
I am using the LO-X7 controller with a Synrad 48-1 tube.
I've found that the TP or Pre-ignition is not present from the X7 controller on TTL or PWM outputs.

Attached are scope comparisons to the Trotec Speedy 100 tickle pulses and the lack of tickle pulses from the 708 controller.

Any help would be greatly appreciated.

Thanks!
ian4goats
Posts: 17
Joined: Wed Apr 09, 2014 10:26 am
Location: Danvers, MA
Contact:

Re: No tickle pulse present - LO-X7

Post by ian4goats »

It's nice to see such good support.
Looks like I'll be dealing directly with Sinjoe.
hodgeac
Posts: 5
Joined: Fri Jun 16, 2017 6:41 pm
Contact:

Re: No tickle pulse present - LO-X7

Post by hodgeac »

ian4goats wrote:It's nice to see such good support.
Looks like I'll be dealing directly with Sinjoe.
I'm actually surprised. It seems like there has been more unanswered posts here lately. Especially considering the threads I read elsewhere saying to spend the extra money and buy from lightobject...specifically for the extra support that comes with buying here. Not impressed.
Tech_Marco
Posts: 4646
Joined: Mon Jun 15, 2009 3:00 pm
Contact:

Re: No tickle pulse present - LO-X7

Post by Tech_Marco »

Sorry I couldn't make support for last 7 days
I have been sick and need time to rest

Btw, I don't have RF tube so there is some question I won't be able to answer


Marco
ian4goats
Posts: 17
Joined: Wed Apr 09, 2014 10:26 am
Location: Danvers, MA
Contact:

Re: No tickle pulse present - LO-X7

Post by ian4goats »

Sorry to hear you're not well Marco.

Do you have an Oscilloscope?

It's easy to check for Tickle pulse or Pre-ionization frequency.
It should be present at .5% duty cycle or 1us at 5Khtz.

Picture is really attached this time.
Trotec Speedy 100 TP.jpg
Tech_Marco
Posts: 4646
Joined: Mon Jun 15, 2009 3:00 pm
Contact:

Re: No tickle pulse present - LO-X7

Post by Tech_Marco »

I talked to the programmer and was told that the feature bus available

Press "Stop + Shift" on the LCD panel to get into a menu where you can set it up on laser properties


Marco
ian4goats
Posts: 17
Joined: Wed Apr 09, 2014 10:26 am
Location: Danvers, MA
Contact:

Re: No tickle pulse present - LO-X7

Post by ian4goats »

Thank you Marco.

This is very good and I wish it was in the manual somewhere.
lawrencejeff
Posts: 3
Joined: Mon Sep 24, 2018 6:26 pm
Contact:

Re: No tickle pulse present - LO-X7

Post by lawrencejeff »

For anyone else's reference - It seems you can only do this with a combination of software and the hardware

In the software (Raycam) set the properties to RF laser (Preignition) with a 5khz frequency.
That by itself doesn't generate the tickle like you would expect, you need to go in the hidden menu (shift+stop) and set the minimum power level to .8
(that was the value for me which generated a synrad compatible tickle 1.5us or so according to my scope)