5740 amp meter

Posted: Mon Feb 28, 2011 11:11 am
by fergusson1
I have a sollar array that is producing about 65 to 75 amps so i bought a 5740 amp meter i followed the instructions hooked for the Case 3 dia. entered each password made sure that each Set point read the way the instructions say they should. The meter reads 499.9 to500.0 this iis not the reading I should be getting so I double checked the instructions going thru each step still get same reading HELP PLEASE

Re: 5740 amp meter

Posted: Tue Mar 01, 2011 11:21 pm
by Tech_Marco
What is specification of the shunt you have? 100A 75mV may be?? Or something else? The setpoint has nothing to do with the reading. It's for the alarm output.

Re: 5740 amp meter

Posted: Wed Mar 02, 2011 7:48 am
by fergusson1
The shunt is a 100 amp 75 mil.

Re: 5740 amp meter

Posted: Thu Mar 03, 2011 6:12 pm
by Tech_Marco
First of all, select Signal type to '3' using password 0089

Try the following setting (display only):
PvL: 0
PvH: 100
dot: 1
Filt: 0

The max display it will show is "100.0"A

Re: 5740 amp meter

Posted: Mon Mar 07, 2011 8:10 am
by fergusson1
isnt this password 36 ?

Re: 5740 amp meter

Posted: Mon Mar 07, 2011 8:33 am
by Tech_Marco
First of all, select Signal type to '3' using password 0089
Then do the other with 0036.
Please follow the manual step by step

Re: 5740 amp meter

Posted: Wed Mar 09, 2011 7:41 am
by fergusson1
Thank you Marco IT WORKS!!!!!!!!!!!!!! one more question I need to mount the shunt can I drill a 1/8 hole in the center of it without ruining the integrity of it ?
Thanks again

Re: 5740 amp meter

Posted: Wed Mar 09, 2011 11:08 am
by Tech_Marco
You can drill the hole on the fixture that holds on the shunt. The size of that pieces of metal (shunt) is precised calculated so don't do any change on it.

Re: 5740 amp meter

Posted: Tue Dec 29, 2015 2:28 pm
by fergusson1
Tech_Marco wrote:You can drill the hole on the fixture that holds on the shunt. The size of that pieces of metal (shunt) is precised calculated so don't do any change on it.
marco
why does this 5740 blink? and how do I fix it

Re: 5740 amp meter

Posted: Tue Dec 29, 2015 2:49 pm
by Tech_Marco
I'm not sure what you mean of blinking on the 5740. Can you video it please

Marco